RIC Morocco

RIC at Casablanca, Morocco (Region I)

(Information on this webpage is based on the latest report from the RIC. 
Please refer to the RIC's website for the latest information.)
General Site Information
Address: Direction de la Météorologie Nationale, Avenue Capitane Tayeb Naciri, administrative complex Hay Hassani, 20220 Casablanca, Morocco
Website: http://cri.marocmeteo.ma
Contact person: Mounir AZIZ
Email: azizmounir@gmail.com
Tel:  +212 522 90 20 08
Fax: +212 522 90 85 93
Calibration capabilities: Temperature, Relative Humidity, Atmospheric Pressure, Precipitation, Solar Radiation
Specific Information on instrument calibration capabilities
Temperature
Instrument Undergoing Calibration
Calibration Range
Reference standard, Equipment
Calibration and Measurement Capability (CMC)*
Traceability of Reference equipment
Last standard calibration date
Calibration body
RTD PT100 -4 wire

-20 to 60 °C

SPRT100 +MicroK precision thermometry bridge
0.065°K
11/12/2020
MCC (accreditation n°MCI/CA/AL34.01/2008)
Status of accreditation (date of the latest accreditation):
Link to the accreditation certificate
Accreditation body:
Relative Humidity
Instrument Undergoing Calibration
Calibration Range
Reference standard, Equipment
Calibration and Measurement Capability (CMC)*
Traceability of Reference equipment
Last standard calibration date
Calibration body

Capacitive hygrometer

 
 
05% to 95%
 
 
 

MBW473 + Hygrogen Generator /HG600 Generator

2*(0.54%+0.006*RH)
 
11/12/2020
MCC (accreditation n°MCI/CA/AL34.01/2008)
Status of accreditation (date of the latest accreditation):
Link to the accreditation certificate
Accreditation body:
Atmospheric Pressure
Instrument Undergoing Calibration
Calibration Range
Reference standard, Equipment
Calibration and Measurement Capability (CMC)*
Traceability of Reference equipment
Last standard calibration date
Calibration body
Digital barometers
500 to 1100 hPa 

CPC6000 pressure module +Automatic generator

6 Pa + 2.1 * 10^-5 P(abs)
22/07/2021
DMT
Electronic barometer
 

500 hPa to 1100 hPa

 

CPC6000 pressure module +Automatic generator

6 Pa + 2.1 * 10^-5 P(abs)
 
Status of accreditation (date of the latest accreditation):
Link to the accreditation certificate
Accreditation body:
Precipitations
Instrument Undergoing Calibration
Calibration Range
Reference standard, Equipment
Calibration and Measurement Capability (CMC)*
Traceability of Reference equipment
Last standard calibration date
Calibration body
Tipping bucket rain gauge

0 to 1000 mm/h

TB340A
±0.3 %F.S. (±2 g)
30/11/2021
HS Hyquest Solutions
Status of accreditation (date of the latest accreditation):
Link to the accreditation certificate
Accreditation body:
Solar Radiation
Instrument Undergoing Calibration
Calibration Range
Reference standard, Equipment
Calibration and Measurement Capability (CMC)*
Traceability of Reference equipment
Last standard calibration date
Calibration body
Pyranometer

Up to 1400 W/m²

SMP21 (Indoor-Kipp & Zonnen Calibration bench)
2%
21/01/2021
Kipp & Zonen
Status of accreditation (date of the latest accreditation):
Link to the accreditation certificate
Accreditation body:

* A CMC (calibration and measurement capability) is the smallest uncertainty of measurement that can be expected to be achieved by the RIC during a calibration. This CMC is evaluated by the RIC itself and described in the scope of accreditation of the RIC, if available.

Further Information

Report of the RIC (2023)

Report of the RIC (2022)

Report of the RIC (February 2020)