RIC at Ljubljana, Slovenia (Region VI)
|
(Information on this webpage is based on the latest report from the RIC.
Please refer to the RIC's website for the latest information.)
|
General Site Information
|
Address: Slovenian Environment Agency, Vojkova 1b, 1000 Ljubljana, Slovenia |
Website: https://www.gov.si/drzavni-organi/organi-v-sestavi/agencija-za-okolje/o-agenciji-zaokolje/ urad-za-okoljska-merjenja/ |
Contact person: Drago GROSELJ |
Email: Drago.Groselj@gov.si |
Tel: +386 1 478 4100 |
Fax: |
Calibration capabilities: Temperature, Relative Humidity, Atmospheric Pressure, Air Quality |
Specific Information on instrument calibration capabilities
|
|
Instrument Undergoing Calibration
|
Calibration Range
|
Reference standard, Equipment
|
Calibration and Measurement Capability (CMC)*
|
Traceability of Reference equipment
|
Last standard calibration date
|
Calibration body
|
Platinum resistance thermometers
|
0.01 °C
|
Triple point of water cell (Isotech, 811L-32)
|
0.7 mK
|
20 July 2020
|
NMI - LMK (Slovenia)
|
Platinum resistance thermometers
|
29.7646 °C
|
Gallium melting point cell (Isotech, ITL-M-17401)
|
0.9 mK
|
20 July 2020
|
NMI - LMK (Slovenia)
|
Resistance thermometers
|
-50 to 50 °C
|
Standard platinum resistance thermometer (Isotech SPRT 419, Fluke 7381, DC bridge MI6010)
|
12 mK
|
05 July 2022
|
NMI - LMK (Slovenia)
|
Liquid-in-glass themrometers
|
-20 to 50 °C
|
Platinum resistance thermometer (Fluke 5626, Fluke 7381, Fluke 1595A)
|
25 mK
|
19 September 2022
|
Slovenian Environment Agency
|
Self-indicated thermometers
|
-50 to 50 °C
|
Platinum resistance thermometer (Fluke 5626, Fluke 7381, Fluke 1595A)
|
25 mK
|
19 September 2022
|
Slovenian Environment Agency
|
|
Status of accreditation (date of the latest accreditation): Valid, accredited according to ISO/IEC 17025:2017, 16 May 2023 |
Link to the accreditation certificate |
Accreditation body: Slovenian Accreditation (SA) |
|
Instrument Undergoing Calibration
|
Calibration Range
|
Reference standard, Equipment
|
Calibration and Measurement Capability (CMC)*
|
Traceability of Reference equipment
|
Last standard calibration date
|
Calibration body
|
Hygrometer
|
10 to 95 %RH
|
Dew-point hygrometer
(MBW 373LHX,Thunder Scientific 2500ST)
|
1.9 %RH (Ta=-10 °C to 0 °C)
1.6 %RH (Ta=0 °C to 10 °C)
1.0 %RH (Ta=10 °C to 50 °C)
|
28 November 2023
|
NMI - LMK (Slovenia)
|
Hygrometer
|
10 to 95 %RH
|
Dew-point meter
(MBW 473, Vötsch VC3 7100)
|
4.2 %RH (Ta=-20 °C to 0 °C)
3.2 %RH (Ta=0 °C to 40 °C)
|
6 February 2024
|
NMI - LMK (Slovenia)
|
|
Status of accreditation (date of the latest accreditation): Valid, accredited according to ISO/IEC 17025:2017, 16 May 2023 |
Link to the accreditation certificate |
Accreditation body: Slovenian Accreditation (SA) |
|
Instrument Undergoing Calibration
|
Calibration Range
|
Reference standard, Equipment
|
Calibration and Measurement Capability (CMC)*
|
Traceability of Reference equipment
|
Last standard calibration date
|
Calibration body
|
Digital barometer
|
200 to 1800 hPa
|
Digital piston gauge
Ruska 2465A-754
(Generator DH type PPC1)
|
3.1 Pa + 1.9·10-5 P
|
9 June 2023
|
CMI (Czech Metrology Institute))
|
|
Status of accreditation (date of the latest accreditation): Valid, accredited according to ISO/IEC 17025:2017, 16 May 2023 |
Link to the accreditation certificate |
Accreditation body: Slovenian Accreditation (SA) |
|
Instrument Undergoing Calibration
|
Calibration Range
|
Reference standard, Equipment
|
Calibration and Measurement Capability (CMC)*
|
Traceability of Reference equipment
|
Last standard calibration date
|
Calibration body
|
Cylinders or generators CO
|
300 to 15000 ppbv
|
Reference analyser, CRMs
|
150 ppbv + 0.025 • c(CO)
|
8 August 2022
|
CHMI (Checz Hydrometeorological Institute)
|
Cylinders or generators SO2
|
3 to 500 ppbv
|
Reference analyser, CRMs
|
2 ppbv + 0.05 • c(SO2)
|
8 June 2023
|
CHMI (Checz Hydrometeorological Institute)
|
Cylinders or generators NO
|
5 to 500 ppbv
|
Reference analyser, CRMs
|
5 ppbv + 0.04 • c(NO)
|
8 June 2023
|
CHMI (Checz Hydrometeorological Institute)
|
Cylinders or generators NO2
|
6 to 500 ppbv
|
Reference analyser, CRMs
|
6 ppbv + 0.04 • c(NO2)
|
8 June 2023
|
CHMI (Checz Hydrometeorological Institute)
|
Analyser CO
|
0 to 16000 ppbv
|
CRMs, zero air generator
|
50 ppbv + 0.026 • c(CO)
|
8 June 2023
|
CHMI (Checz Hydrometeorological Institute)
|
Analyser SO2
|
0 to 500 ppbv
|
CRMs, zero air generator
|
0.7 ppbv + 0.05 • c(SO2)
|
8 June 2023
|
CHMI (Checz Hydrometeorological Institute)
|
Analyser NO (NOx)
|
0 to 500 ppbv
|
CRMs, zero air generator
|
3.5 ppbv + 0.04 • c(NO)
|
8 June 2023
|
CHMI (Checz Hydrometeorological Institute)
|
Analyser O3
|
0 to 500 ppbv
|
CRMs, zero air generator
|
2 ppbv + 0.035 • c(O3)
|
8 June 2023
|
CHMI (Checz Hydrometeorological Institute)
|
|
Status of accreditation (date of the latest accreditation): Valid, accredited according to ISO/IEC 17025:2017, 16 May 2023 |
Link to the accreditation certificate |
Accreditation body: Slovenian Accreditation (SA) |