RIC Germany

RIC at Hamburg/Oberschleissheim, Germany (Region VI)

(Information on this webpage is based on the latest report from the RIC.
Please refer to the RIC's website for the latest information.)
General Site Information
Address: Frahmredder 95, 22393 Hamburg, Hamburg,Germany
Websitehttps://www.dwd.de/RIC
Contact person: Holger DOERSCHEL
Email: holger.doerschel@dwd.de
Tel: +49 6980626688
Fax: +49 6980626699
Calibration capabilities: Temperature, Relative Humidity, Atmospheric Pressure, Wind, Precipitation
Specific Information on instrument calibration capabilities
Temperature
Instrument Undergoing Calibration
Calibration Range
Reference standard, Equipment
Calibration and Measurement Capability (CMC)*
Traceability of Reference equipment
Last standard calibration date
Calibration body

Pt100 4-Wire(Hamburg Setup 1/2)

-30 to 45 °C

Fluke 1594A Super-ThermomterFluke 2590 Multiplexer, Fluke 7341

Klasmeier eXacal Pt100M

0.02 K

11/2019

06/2019

DAkKS

DAkkS

Pt100 4-Wire(Hamburg Setup 2/2) -30 to 45 °C

Fluke 1594A Super-ThermomterFluke 2590 Multiplexer, Fluke 7341

Klasmeier eXacal Pt100M

0.02 K

11/2019

07/2019

DAkKS

DAkkS

 

Elektronic Thermometers(Hamburg)

-30 to 45 °C

MBW T12 CTS T -70/200

0.06 K

 01/2020

SCS

Pt100 4-Wire(Oberschleissheim) -30 to 45 °C

Fluke 1594A Super-ThermomterFluke 2590 Multiplexer, Fluke 7341

SPRT Nr. 1

SPRT Nr. 2

0.02 K

11/2019

08/2019

03/2018

DAkkS

DAkkS

DAkkS

Elektronic Thermometers(Oberschleissheim) -30 to 45 °C

MBW T12 Weiss WK3-600/40-S

0.06 K 12/2019 SCS

Status of accreditation (date of the latest accreditation): Valid, accredited 12/2020 according to ISO/IEC 17025:2017

Link to the accreditation certificate:
Accreditation body:
Relative Humidity
Instrument Undergoing Calibration
Calibration Range
Reference standard, Equipment
Calibration and Measurement Capability (CMC)*
Traceability of Reference equipment
Last standard calibration date
Calibration body

Electronic Hygrometers(Hamburg Setup 1/3)

15 to 95 %RH
(at 20 to 25 °C)

MBW DPM473 NR. 1 Thunder 2500 Nr. 1

0.01% + 0.006%*RH,

but not less than
0.4 %RH

 

02/2019

 

PTB

Electronic Hygrometers(Hamburg Setup 2/3) 15 to 95 %RH
(at 20 to 25 °C)
MBW DPM473 Nr. 2 Thunder 2500 Nr. 2

0.01% + 0.006%*RH,

but not less than
0.4 %RH

 

 

03/2020

 

PTB

Electronic Hygrometers(Hamburg Setup 3/3)   MBW DPM473 Nr. 3 Thunder 2500 Nr. 3

0.01% + 0.006%*RH,

but not less than
0.4 %RH

 

02/2020

 

SCS

Electronic Hygrometers(Oberschleissheim Setup 1/2) 15 to 95 %RH
(at 20 to 25 °C)
MBW DPM473 NR. 1 Thunder 2500 Nr. 1

0.01% + 0.006%*RH,

but not less than
0.4 %RH

 

02/2019

 

PTB

Electronic Hygrometers(Oberschleissheim Setup 2/2) 15 to 95 %RH
(at 20 to 25 °C)
MBW DPM473 Nr. 2 Thunder 2500 Nr. 2

0.01% + 0.006%*RH,

but not less than
0.4 %RH

 

03/2020

 

PTB

Status of accreditation (date of the latest accreditation): Valid, accredited 12/2020 according to ISO/IEC 17025:2017

Link to the accreditation certificate
Accreditation body:
Pressure
Instrument Undergoing Calibration
Calibration Range
Reference standard, Equipment
Calibration and Measurement Capability (CMC)*
Traceability of Reference equipment
Last standard calibration date
Calibration body

Electronic Barometers(Hamburg)

100 to 2500 hPa

DHI 7601 Piston GaugeAMH38 Automatic Mass Handler

Inficon CG025D Vacuum meter

2.0*10-5 * Pabs + 0.0025 hPa

05/2019

06/2019

PTB

 

PTB

Electronic Barometers(Oberschleissheim) 100 to 2500 hPa

DHI 7601 Piston GaugeAMH38 Automatic Mass Handler

Inficon CG025D Vacuum meter

2.0*10-5 * Pabs + 0.0025 hPa

06/2015

06/2019

PTB

 

PTB

Status of accreditation (date of the latest accreditation): Valid, accredited 12/2020 according to ISO/IEC 17025:2017
Link to the accreditation certificate

Accreditation body: 

Wind direction
Instrument Undergoing Calibration
Calibration Range
Reference standard, Equipment
Calibration and Measurement Capability (CMC)*
Traceability of Reference equipment
Last standard calibration date
Calibration body

Wind speed instruments (Hamburg)

0 to 360°

SICK AFS60A

Windtunnel Goettingen Type

0,9°

 

04/2019

 

DAkkS

Wind speed instruments (Oberschleissheim)

0 to 360°

SICK AFS60A

Windtunnel Goettingen Type

0,9°

09/2019

DAkkS

Status of accreditation (date of the latest accreditation): Valid, accredited 12/2020 according to ISO/IEC 17025:2017
Link to the accreditation certificate

Accreditation body: 

Wind speed
Instrument Undergoing Calibration
Calibration Range
Reference standard, Equipment
Calibration and Measurement Capability (CMC)*
Traceability of Reference equipment
Last standard calibration date
Calibration body

Wind speed instruments (Hamburg)

0.15 to 50 m/s

Dantec 2D LDA

Windtunnel Goettingen Type

1,0 % but not less than 0,1 m/s

 

10/2018

 

PTB

Wind speed instruments (Oberschleissheim)

0.15 to 50 m/s

Dantec 2D LDA

Windtunnel Goettingen Type

1,0 % but not less than 0,1 m/s

11/2019

PTB

Status of accreditation (date of the latest accreditation): Valid, accredited 12/2020 according to ISO/IEC 17025:2017
Link to the accreditation certificate

Accreditation body

 

Precipitation
Instrument Undergoing Calibration
Calibration Range
Reference standard, Equipment
Calibration and Measurement Capability (CMC)*
Traceability of Reference equipment
Last standard calibration date
Calibration body

Weighing/tipping bucketinstruments (Hamburg)

0.1 to 3.0 mm/mm

Kern PBS6200-2M

2%, not less than 0,2 mm/min

 

07/2019

 

SCS

Weighing/tipping bucketInstruments (Oberschleissheim) 0.1 to 3.0 mm/mm Kern PBS6200-2M 2 %, not less than 0,2 mm

07/2019

SCS

Status of accreditation (date of the latest accreditation): Valid, accredited 12/2020 according to ISO/IEC 17025:2017
Link to the accreditation certificate

Accreditation body

SCS - Swiss Calibration Service

DAkKS - Deutsche Akkreditierungsstelle GmbH / German Accreditation Body

PTB - Physikalisch-Technische Bundesanstalt / The National Metrology Institute of Germany

A CMC (calibration and measurement capability) is the smallest uncertainty (k=2) of measurement that can be expected to be achieved by the RIC during a calibration under normal conditions. This CMC is evaluated by the RIC itself and described in the scope of accreditation of the RIC, if available.

Further Information

 Report of the RIC (January 2021)

Report of the RIC (February 2020)